Reproducing Simulated Chromatograms from "Predictive mechanistic modeling of loading and elution in protein A chromatography" article from Lenhoff (2024)
Hello community,
I am new to CADET modeling and still gaining familiarity with chromatography process modeling. I am currently working on reproducing the simulated chromatograms presented in: “Bhoyar, S.; Kumar, V.; Foster, M.; Xu, X.; Traylor, S.; Guo, J.; Lenhoff, A.: Predictive mechanistic modeling of loading and elution in protein A chromatography, Journal of Chromatography A 1713 (2024): 464558” Specifically, I am attempting to replicate the chromatograms shown in Figures 10a, 10b, and 10c (Section 3.2.2) using CADET-Python. I implemented the reported model architecture, unit operations, solver settings, and isotherm parameters as described in the article. However, I had to make two assumptions due to missing details:
  1. pH Inlet Profile Construction:
    I manually parameterized the inlet pH trajectory based on Section 2.2.3 of the paper, i.e., constant pH 7.5 during load and wash, followed by a linear gradient to pH 3.5 over 25 CV for elution.
  2. Loading Duration Estimation:
    As the manuscript does not explicitly state the loading duration for the simulations, I inferred these values from the experimental chromatograms shown (e.g., ~115 CV for Figure 10b).
While my simulated loading curves show small differences compared to the paper, the deviations during the wash and particularly elution steps are more pronounced.
Key discrepancies include:
a) Significant loss of mAb during the wash phase, which is not observed in the published simulations
b) Earlier onset of elution
c) Lower elution peak height
d) Outlet pH at peak maximum deviates from the published curves
e) Overall chromatogram shape differs from the solid black simulation curves in the paper
For reference, I have attached representative plots (blue = chromatogram, red = outlet pH) below. I have also included my full CADET-Python script at the end of this post.
  1. from the article (fig 10 a–c)
  2. from my code
    Given the differences described above, I would greatly appreciate insight from the community. Any suggestions, corrections, or perspectives would be extremely valuable.
Thank you in advance for your time and help.
Code:
#!/usr/bin/env python3
Standalone reproduction script for a 1x3 chromatogram panel (L=2.5 cm, c0=6 mg/mL)
with outlet pH on a secondary y-axis and top x-axis ticks (no labels).
Runs CADET if H5 files are missing.
import os
import math
import h5py
try:
except Exception as exc:
    raise RuntimeError("cadet-python is required to run this script") from exc
OUTPUT_DIR = os.path.join(os.path.dirname(__file__), "out")
OUTPUT_PNG = os.path.join(OUTPUT_DIR, "chromatograms_L2.5_c6_v75_v150_v300_panel_repro.png")
# Column and system
ID = 0.03
R = ID / 2.0
EPS_C = 0.40
EPS_P = 0.65
R_P = 44e-6
LENGTH_CM = 2.5
# Transport
D_AX = 1.0e-6
K_FILM = 1.0e-5
D_P = 4.0e-12
D_H = 1e-9
# Colloidal constants
A_PROT = 4.5e-9
PHI = 1.6e8
KAPPA_MAB4 = 5.3e-9
# Discretization
NCOL = 60
NPAR = 45
ABS_TOL = 1e-8
REL_TOL = 1e-6
N_TIME_POINTS = 2000
SENSITIVITY_TIME_POINTS = 500
MW_MAB = 150000.0
SURF_DIFF_Hp = 0.0
SURF_DIFF_PROTEIN = 4.6e-14
SURF_DIFF_DEP = "LIQUID_SALT_EXPONENTIAL"
SURF_DIFF_EXP_FACTOR_PROTEIN = 1.0
SURF_DIFF_EXP_ARGMULT_PROTEIN = -26800.0
PH_LOAD = 7.5
PH_END = 3.5
N_ELUT_SEGMENTS = 80
LOAD_CV_DEFAULT = 115.0
LOAD_CV_75 = 60.0
WASH_CV = 60.0
ELUT_CV = 30.0
VELOCITIES_CM_PER_H = [75.0, 150.0, 300.0]
C0_MG_ML = 6.0
RUN_SENSITIVITY = True
SENSITIVITY_REQUIRE_EXISTING = False
def mg_per_mL_to_mol_per_m3(c_mg_per_mL: float, mw_g_per_mol: float) -> float:
    c_g_per_L = c_mg_per_mL
    c_mol_per_L = c_g_per_L / mw_g_per_mol
    return c_mol_per_L * 1000.0
def cv_to_time_s(cv_value: float, length_cm: float, velocity_cm_per_h: float) -> float:
    u_s = (velocity_cm_per_h / 100.0) / 3600.0
    t_per_CV = (length_cm / 100.0) / u_s if u_s else None
    return float(cv_value) * t_per_CV if t_per_CV else None
def build_and_run(
    length_cm,
    velocity_cm_per_h,
    c0_mg_ml,
    h5_name,
    transport_overrides=None,
    n_time_points=None,
    load_cv = LOAD_CV_75 if abs(velocity_cm_per_h - 75.0) < 1e-9 else LOAD_CV_DEFAULT
    t_load = cv_to_time_s(load_cv, length_cm, velocity_cm_per_h)
    t_wash = cv_to_time_s(WASH_CV, length_cm, velocity_cm_per_h)
    t_elut = cv_to_time_s(ELUT_CV, length_cm, velocity_cm_per_h)
    t_total = t_load + t_wash + t_elut
    c0 = mg_per_mL_to_mol_per_m3(c0_mg_ml, MW_MAB)
    H_load = 10.0 ** (-PH_LOAD) * 1000.0
    m = Cadet()
    m.filename = os.path.abspath(os.path.join(OUTPUT_DIR, h5_name))
    os.makedirs(os.path.dirname(m.filename), exist_ok=True)
    ncomp = 2
    m.root.input.model.nunits = 3
    m.root.input.model.unit_000.unit_type = "INLET"
    m.root.input.model.unit_000.ncomp = ncomp
    m.root.input.model.unit_000.inlet_type = "PIECEWISE_CUBIC_POLY"
    elut_dt = t_elut / N_ELUT_SEGMENTS if N_ELUT_SEGMENTS > 0 else t_elut
    section_times = [0.0, t_load, t_load + t_wash]
    for i in range(N_ELUT_SEGMENTS):
        section_times.append(t_load + t_wash + (i + 1) * elut_dt)
    m.root.input.solver.sections.nsec = len(section_times) - 1
    m.root.input.solver.sections.section_times = section_times
    m.root.input.solver.sections.section_continuity = [0] * (len(section_times) - 2)
    m.root.input.model.unit_000.sec_000.const_coeff = [H_load, c0]
    m.root.input.model.unit_000.sec_000.lin_coeff = [0.0, 0.0]
    m.root.input.model.unit_000.sec_000.quad_coeff = [0.0, 0.0]
    m.root.input.model.unit_000.sec_000.cube_coeff = [0.0, 0.0]
    m.root.input.model.unit_000.sec_001.const_coeff = [H_load, 0.0]
    m.root.input.model.unit_000.sec_001.lin_coeff = [0.0, 0.0]
    m.root.input.model.unit_000.sec_001.quad_coeff = [0.0, 0.0]
    m.root.input.model.unit_000.sec_001.cube_coeff = [0.0, 0.0]
# We approximate the H+ profile for elution section by splitting elution into many
# small segments, because H+ concentration is defined in terms of inlet pH: linear pH
# becomes a nonlinear H+ profile. Piecewise segments approximate the intended H+ profile.
        seg_start = t_load + t_wash + i * elut_dt
        seg_end = seg_start + elut_dt
        frac0 = (seg_start - (t_load + t_wash)) / t_elut if t_elut > 0 else 0.0
        frac1 = (seg_end - (t_load + t_wash)) / t_elut if t_elut > 0 else 1.0
        frac0 = min(max(frac0, 0.0), 1.0)
        frac1 = min(max(frac1, 0.0), 1.0)
        pH0 = PH_LOAD + (PH_END - PH_LOAD) * frac0
        pH1 = PH_LOAD + (PH_END - PH_LOAD) * frac1
        H0 = 10.0 ** (-pH0) * 1000.0
        H1 = 10.0 ** (-pH1) * 1000.0
        slope = (H1 - H0) / elut_dt if elut_dt > 0 else 0.0
        sec_idx = 2 + i
        sec = getattr(m.root.input.model.unit_000, f"sec_{sec_idx:03d}")
        sec.const_coeff = [H0, 0.0]
        sec.lin_coeff = [slope, 0.0]
        sec.quad_coeff = [0.0, 0.0]
        sec.cube_coeff = [0.0, 0.0]
    col = m.root.input.model.unit_001
    col.unit_type = "GENERAL_RATE_MODEL"
    col.ncomp = ncomp
    col.col_length = length_cm / 100.0
    col.col_radius = R
    col.col_porosity = EPS_C
    col.par_porosity = EPS_P
    col.par_radius = R_P
    col.col_dispersion = D_AX
    col.velocity = (velocity_cm_per_h / 100.0) / 3600.0
    col.film_diffusion = [K_FILM, K_FILM]
    col.par_diffusion = [D_H, D_P]
    col.PAR_SURFDIFFUSION = [SURF_DIFF_Hp, SURF_DIFF_PROTEIN]
    col.PAR_SURFDIFFUSION_DEP = SURF_DIFF_DEP
    if SURF_DIFF_DEP == "LIQUID_SALT_EXPONENTIAL":
        col.PAR_SURFDIFFUSION_EXPFACTOR = [0.0, SURF_DIFF_EXP_FACTOR_PROTEIN]
        col.PAR_SURFDIFFUSION_EXPARGMULT = [0.0, SURF_DIFF_EXP_ARGMULT_PROTEIN]
    col.discretization.ncol = NCOL
    col.discretization.npar = NPAR
    col.discretization.par_disc_type = "EQUIDISTANT"
    col.discretization.use_analytic_jacobian = 1
    col.discretization.use_implicit_volume = 1
    col.discretization.gs_type = 1
    col.discretization.max_krylov = 0
    col.discretization.max_restarts = 10
    col.discretization.schur_safety = 1e-8
    col.discretization.use_weno = 0
    col.discretization.weno.boundary_model = 0
    col.discretization.weno.weno_eps = 1e-6
    col.discretization.weno.weno_order = 3
    col.adsorption_model = "MULTI_COMPONENT_COLLOIDAL"
    col.adsorption.COL_PHI = PHI
    col.adsorption.COL_CORDNUM = 6
    col.adsorption.IS_KINETIC = 1
    col.adsorption.COL_PROTEIN_RADIUS = [0, A_PROT]
    col.adsorption.COL_USE_PH = 0
    col.adsorption.COL_LINEAR_THRESHOLD = 1e-8
    col.adsorption.COL_LOGKEQ_SALT_POWEXP = [0, 0.0133]
    col.adsorption.COL_LOGKEQ_SALT_POWFACT = [0, 0.38]
    col.adsorption.COL_LOGKEQ_SALT_EXPFACT = [0, 19.47]
    col.adsorption.COL_LOGKEQ_SALT_EXPARGMULT = [0, -16000.0]
    col.adsorption.COL_LOGKEQ_PH_EXP = [0, 0]
    col.adsorption.COL_BPP_SALT_POWFACT = [0, 0.0004]
    col.adsorption.COL_BPP_SALT_POWEXP = [0, -0.516]
    col.adsorption.COL_BPP_SALT_EXPFACT = [0, 9.348]
    col.adsorption.COL_BPP_SALT_EXPARGMULT = [0, -6993.0]
    col.adsorption.COL_BPP_PH_EXP = [0, 0]
    col.adsorption.COL_KKIN = [1e9, 1e9]
    col.adsorption.COL_KAPPA_EXP = 0
    col.adsorption.COL_KAPPA_FACT = 0
    col.adsorption.COL_KAPPA_CONST = 1 / KAPPA_MAB4
    if transport_overrides:
        for key, value in transport_overrides.items():
            setattr(col, key, value)
    col.nbound = [0, 1]
    col.init_c = [H_load, 0.0]
    col.init_q = [0.0, 0.0]
    m.root.input.model.unit_002.unit_type = "OUTLET"
    m.root.input.model.unit_002.ncomp = ncomp
    Q = col.velocity * math.pi * R * R
    m.root.input.model.connections.nconn = 2
    m.root.input.model.connections.nswitches = 1
    m.root.input.model.connections.switch_000.section = 0
    m.root.input.model.connections.switch_000.connections = [0, 1, -1, -1, Q, 1, 2, -1, -1, Q]
    rgrp = m.root.input["return"]
    rgrp.WRITE_SOLUTION_TIMES = 1
    rgrp.SPLIT_COMPONENTS_DATA = 1
    rgrp.SPLIT_PORTS_DATA = 1
    rgrp.unit_001.WRITE_SOLUTION_OUTLET = 1
    rgrp.unit_002.WRITE_SOLUTION_OUTLET = 1
    m.root.input.solver.time_integrator.abstol = ABS_TOL
    m.root.input.solver.time_integrator.reltol = REL_TOL
    m.root.input.solver.time_integrator.algtol = 1e-10
    m.root.input.solver.time_integrator.init_step_size = 1e-6
    m.root.input.solver.time_integrator.max_steps = 1000000
    m.root.input.model.solver.schur_safety = 1e-8
    m.root.input.model.solver.gs_type = 1
    m.root.input.model.solver.max_krylov = 0
    m.root.input.model.solver.max_restarts = 10
    time_points = N_TIME_POINTS if n_time_points is None else n_time_points
    if time_points and t_total > 0:
        m.root.input.solver.user_solution_times = np.linspace(0.0, t_total, time_points)
    m.save()
    ret = m.run_simulation()
    if getattr(ret, "return_code", 1) not in (0, 1):
        stderr = getattr(ret, "stderr", None)
        stdout = getattr(ret, "stdout", None)
        msg = f"CADET run failed (code={getattr(ret, 'return_code', None)})"
        if stderr:
            msg += f"\nSTDERR:\n{stderr}"
        if stdout:
            msg += f"\nSTDOUT:\n{stdout}"
        raise RuntimeError(msg)
    return m.filename
def read_outlet_comp(h5_path: str, comp_idx: int):
    with h5py.File(h5_path, "r") as fh:
        sol = fh.get("/output/solution") or fh.get("output/solution")
        if sol is None:
            return None, None
        times = sol.get("SOLUTION_TIMES")[()]
        unit = sol.get("unit_002") or sol.get("unit_001")
        if unit is None:
            return None, None
        key = f"SOLUTION_OUTLET_COMP_{comp_idx:03d}"
        if key in unit:
            comp = unit[key][()]
            comp = None
            for k in unit:
                if "SOLUTION_OUTLET" in k.upper():
                    d = unit[k][()]
                    comp = d[:, comp_idx] if d.ndim == 2 else d
                    break
        return times, comp
def main():
    import matplotlib
    matplotlib.use("Agg")
    os.makedirs(OUTPUT_DIR, exist_ok=True)
    fig, axes = plt.subplots(1, 3, figsize=(14, 4), sharey=False)
    for ax, v_cm_per_h in zip(axes, VELOCITIES_CM_PER_H):
        h5_name = f"L{LENGTH_CM:.1f}_u{v_cm_per_h:.0f}_c{C0_MG_ML:.1f}.h5"
        h5_path = os.path.join(OUTPUT_DIR, h5_name)
        if not os.path.exists(h5_path):
            h5_path = build_and_run(LENGTH_CM, v_cm_per_h, C0_MG_ML, h5_name)
        times, comp1 = read_outlet_comp(h5_path, 1)
        _, comp0 = read_outlet_comp(h5_path, 0)
        if times is None or comp1 is None or comp0 is None:
            if os.path.exists(h5_path):
                os.remove(h5_path)
                raise RuntimeError(f"Missing outlet data in {h5_path}")
        u_s = (v_cm_per_h / 100.0) / 3600.0
        t_per_CV = (LENGTH_CM / 100.0) / u_s
        CV = times / t_per_CV
        c0 = mg_per_mL_to_mol_per_m3(C0_MG_ML, MW_MAB)
        c_over_c0 = comp1 / c0
        ax.plot(CV, c_over_c0, color="#4C78A8", lw=2)
        ax.set_xlabel("Volume (CV)")
        ax.set_ylabel("c/c0")
        x_max = float(np.nanmax(CV)) if np.size(CV) else 0.0
        ax.set_xlim(0, x_max)
        y_max = max(1.0, np.nanmax(c_over_c0) * 1.05)
        y_max = math.ceil(y_max / 0.2) * 0.2
        ax.set_ylim(0, y_max)
        ax.text(0.02, 0.95, f"v = {v_cm_per_h:.0f} cm/h", transform=ax.transAxes, va="top", ha="left")
        ax.set_xticks(np.arange(0, x_max + 1e-6, 50))
        ax.set_xticks(np.arange(0, x_max + 1e-6, 10), minor=True)
        ax.set_yticks(np.arange(0, y_max + 1e-9, 0.2))
        ax.set_yticks(np.arange(0, y_max + 1e-9, 0.05), minor=True)
        ax.tick_params(axis="x", which="minor", length=3)
        ax.tick_params(axis="y", which="minor", length=3)
        ax.grid(True, which="major", alpha=0.3)
        outlet_pH = np.array([(-math.log10(v / 1000.0) if v > 0 else np.nan) for v in comp0])
        ax2 = ax.twinx()
        ax2.plot(CV, outlet_pH, color="#E45756", lw=1.5)
        ax2.set_ylabel("pH")
        pH_min = float(np.nanmin(outlet_pH)) if np.size(outlet_pH) else 0.0
        pH_max = float(np.nanmax(outlet_pH)) if np.size(outlet_pH) else 1.0
        if pH_max <= pH_min:
            pH_min, pH_max = 0.0, 1.0
        pH_min_tick = math.floor(pH_min * 2.0) / 2.0
        pH_max_tick = math.ceil(pH_max * 2.0) / 2.0
        ax2.set_yticks(np.arange(pH_min_tick, pH_max_tick + 1e-9, 0.5))
        ax2.set_yticks(np.arange(pH_min_tick, pH_max_tick + 1e-9, 0.25), minor=True)
        ax2.tick_params(axis="y", which="minor", length=3)
        ax_top = ax.secondary_xaxis("top")
        ax_top.set_xticks(np.arange(0, x_max + 1e-6, 50))
        ax_top.set_xticks(np.arange(0, x_max + 1e-6, 10), minor=True)
        ax_top.set_xlabel("")
        ax_top.tick_params(axis="x", which="major", labeltop=False)
    fig.tight_layout()
    fig.savefig(OUTPUT_PNG, dpi=200)
    plt.close(fig)
    print(f"[INFO] Saved panel: {OUTPUT_PNG}")
if __name__ == "__main__":
    main()
Hi Shubham,
It looks to me like you’re supplying the [H+] in mM, but the parameter values presented in Table 2 in the paper correspond to the [H+] concentration in mol / L, not mmol / L or mol / m^{3}. I’m not sure if the units used were explicitly stated in the paper, but you can tell buy plotting the pH dependent parameter functions using the provided parameter values with [H+] in either mol / L or mmol / L. Comparing these to the data shown in Figure 5 a) and b), the pH dependence functions match up with the results from the paper when [H+] is provided in mol / L. Providing the inlet [H+] profile in mmol / L without adjusting the parameter values would cause both the early elution and the loss during the wash phase that you are seeing.
Since experimental or simulated pH traces that were not perfectly linear were used for the inlet profiles in the paper, I would expect to see some differences in the shapes of the elution peaks if you are using perfectly linear inlet pH profiles. However, I believe accounting for the difference in [H+] units will get you much closer to the expected result.
Hi Angela,
Thank you so much for that insight! That makes sense and indeed I have been supplying [H+] in mM. I have rectified it and the biggest change I noticed that the elution curve is shifted towards the right. Below is the comparison between my old run and the updated one after accounting [H+] conc in M.
The pH trace (shown by dashed lines) also makes sense as supplying [H+] in mol/m3 or (mM) meant pH was actually lower than what is intended, and thus all the parameters (functions) dependent on [H+] would be different as your plots also point out. Thus, I also notice the changes in the loading section (where the slopes are different) and also the elution. (Note: in my first post, I realized that the plotted pH traces (for which I use an external function) did not represent what CADET was considering as while plotting pH traces, I had correctly accounted for the units (converting [H+] to M from mM before taking the logarithm))
However, the updated changes after fixing [H+] unit to M or (mol/L) and supplying to the model still do not match the expected results. If we leave out the elution profile (Which I agree depends on the gradient used in the inlet pH profile), the loading and the wash profiles still deviate. My expectation is that at least these should be very close to that reported in the paper since the inlet pH do not vary in these sections (inlet pH is 7.5 for loading and wash sections).
The above I have shown two chromatograms, the top one is from my simulations, and the bottom one is from the paper (fig 10b). Since the paper does not explicitly suggest the inlet conc. of mAb (gives a range of between 5 - 6 mg/mL or (g/L), in my first post I had considered 6 mg/mL), the two curves in my generated plots are for 5 and 6 mg/mL. Key deviations in the loading and wash sections include:
  1. CV at 99% breakthrough from my simulations is ~10 CV higher than that observed from the simulations in the paper (in paper it looks to be around 40 CV whereas my simulations show 99% breakthrough happening around 50 CV, exact values labelled in the plots)
  2. I still observe more proteins being washed out in my simulations as compared to that observed in the paper. As an illustration, I have drawn a line from c/c0 = 0.05, to check at what CV in the wash section does c/c0 falls below it. My simulations show that c/c0 goes under 0.05 at approximately 25 - 30 CVs later than what is observed in the paper (in paper it seems like c/c0 goes below 0.05 at approximately 125 CV whereas my simulations show that happening at > 150 CV, exact values labelled in the plots)
At this point, I am trying to understand whether the remaining deviations may stem from model assumptions, missing parameters in the paper, or misinterpretations on my end.
If you have any thoughts on what might be contributing to the slower breakthrough or the extended tailing during the wash, I would sincerely appreciate it. Thank you again for your time and guidance!
Shubham