An example to use ACT isotherm for protein A affinity chromatography
Introduction
The ACT isotherm is a modified Langmuir isotherm tailored for mechanistic modeling in protein A chromatography. It is known that both binding capacity q_{max} and equilibrium constant (affinity) K_{eq} changes as a function of pH. ACT isotherm uses modification factors derived based on proA and mAb interaction stoichiometries. Details can be found in the paper.
Example code
Cadet.cadet_path = r"C:\Users\zwend\CADET\cadet_act2025\bin\cadet-cli.exe"     ## IMPORTANT: use a version higher than 5.0.4
## system settings
column_length = 0.025              ## m
column_volume = 5.025e-6           ## m^3
protein_MW = 150                   ## kDa
column_porosity = 0.31             ## 1
particle_porosity = 0.95           ## 1
Q = 1.25 /(6e7)                    ## volumetric flow rate, m^3/s
elution_pH_start = 5.5
elution_pH_end = 3.3
loading = 50.0                     ## mg/mL resin
c_feed = 4.0 / protein_MW          ## mol/m^3
RT = column_volume / Q             ## s
event_CV = [0.0, loading / (c_feed * protein_MW), 4.0, 10.0, ]           ## load, wash, elution in CV
Keq = 268        ## ml/mg
Q_max = 85.6                                                             ## mg/ml column volume
total_porosity = column_porosity + (1.0-column_porosity)*particle_porosity
model = Cadet()
model.root.input.model.unit_000.ncomp = 2                            ## The first component is pH
model.root.input.model.unit_001.unit_type = 'GENERAL_RATE_MODEL'
model.root.input.model.unit_001.ncomp = 2
## Geometry
model.root.input.model.unit_001.col_length = column_length                         # m
model.root.input.model.unit_001.cross_section_area = column_volume/column_length   # m^2
model.root.input.model.unit_001.col_porosity = column_porosity                     # 1
model.root.input.model.unit_001.par_porosity = particle_porosity                   # 1
model.root.input.model.unit_001.par_radius = 0.0425e-3                             # m
## Transport
model.root.input.model.unit_001.col_dispersion = 1.36e-8            # m^2/s
model.root.input.model.unit_001.film_diffusion = [1, 1.41e-5,]      # m/s
model.root.input.model.unit_001.par_diffusion = [1, 1.99e-11]       # m^2/s
model.root.input.model.unit_001.par_surfdiffusion = [0.0, 0.0]
model.root.input.model.unit_001.adsorption_model = 'AFFINITY_COMPLEX_TITRATION'
model.root.input.model.unit_001.adsorption.is_kinetic = 1
model.root.input.model.unit_001.adsorption.act_ka = [1.0, Keq*protein_MW*(1.0-total_porosity), ]          ##  m^3 solid phase / mol protein / s   ml/mg=m^3/kg    1 kda = 1 kg/mol
model.root.input.model.unit_001.adsorption.act_kd = [1.0, 1.0]                                            ## s^-1
model.root.input.model.unit_001.adsorption.act_qmax = [1e-10, Q_max/protein_MW/(1.0-total_porosity),]     ##  mol/m^3 solid phase mg/ml=kg/m^3 / 150 kg/mol = 1/150 mol/m^3
model.root.input.model.unit_001.adsorption.act_etaA = [0, 2.12,  ]
model.root.input.model.unit_001.adsorption.act_pkaA = [0, 3.82, ]
model.root.input.model.unit_001.adsorption.act_etaG = [0, 3.48, ]
model.root.input.model.unit_001.adsorption.act_pkaG = [0, 4.66, ]
model.root.input.model.unit_001.init_c = [7.2, 0.0, ]
model.root.input.model.unit_001.init_q = [0.0, 0.0]
### Grid cells
model.root.input.model.unit_001.discretization.ncol = 50
model.root.input.model.unit_001.discretization.npar = 12
### Bound states
model.root.input.model.unit_001.discretization.nbound = [0, 1,]
### Other options
model.root.input.model.unit_001.discretization.weno.weno_order = 2
model.root.input.model.unit_002.ncomp = 2
model.root.input.solver.sections.nsec = 4
model.root.input.solver.sections.section_times = [0, event_CV[1] * RT, (event_CV[1] + event_CV[2]) * RT, (event_CV[1] + event_CV[2] + event_CV[3]) * RT, (event_CV[1] + event_CV[2] + event_CV[3] + 5.5) * RT]  ## s
model.root.input.solver.sections.section_continuity = [0,0,0,0]
## load
model.root.input.model.unit_000.sec_000.const_coeff = [7.2, c_feed,]                   # mol / m^3
model.root.input.model.unit_000.sec_000.lin_coeff =  [0.0, 0.0,]
model.root.input.model.unit_000.sec_000.quad_coeff = [0.0, 0.0,]
model.root.input.model.unit_000.sec_000.cube_coeff = [0.0, 0.0,]
## wash
model.root.input.model.unit_000.sec_001.const_coeff = [5.5, 0.0,]
model.root.input.model.unit_000.sec_001.lin_coeff =  [0.0, 0.0,]
model.root.input.model.unit_000.sec_001.quad_coeff = [0.0, 0.0,]
model.root.input.model.unit_000.sec_001.cube_coeff = [0.0, 0.0,]
## elution
model.root.input.model.unit_000.sec_002.const_coeff = [elution_pH_start, 0.0]
model.root.input.model.unit_000.sec_002.lin_coeff = [-(elution_pH_start-elution_pH_end)/(event_CV[3] * RT), 0.0, ]
model.root.input.model.unit_000.sec_002.quad_coeff = [0.0, 0.0,]
model.root.input.model.unit_000.sec_002.cube_coeff = [0.0, 0.0,]
## end
model.root.input.model.unit_000.sec_003.const_coeff = [elution_pH_end, 0.0]
model.root.input.model.unit_000.sec_003.lin_coeff = [0.0,0.0,]
model.root.input.model.unit_000.sec_003.quad_coeff = [0.0,0.0,]
model.root.input.model.unit_000.sec_003.cube_coeff = [0.0,0.0,]
    0, 1, -1, -1, Q,
    1, 2, -1, -1, Q]
# Return data
model.root.input['return'].split_components_data = 1
model.root.input['return'].unit_000.write_solution_bulk = 0
model.root.input['return'].unit_000.write_solution_inlet = 0
# Solution times
model.root.input.solver.user_solution_times = np.linspace(0, (event_CV[1] + event_CV[2] + event_CV[3] + 5.5) * RT, 201)
model.filename = 'model.h5'
model.save()
data = model.run()
model.load()
c = model.root.output.solution.unit_001.solution_outlet_comp_001
pH_outlet = model.root.output.solution.unit_001.solution_outlet_comp_000
fig = plt.figure()
ax = fig.add_subplot()
ax.plot(time[1:]/60*Q*6e7, c[1:]*protein_MW, c="orange", label="mAb")
ax.set_xlabel(r'Volume/mL')
ax.set_ylabel(r'Concentration/(mg/mL)')
ax_ph = ax.twinx()
ax_ph.plot(time/60*Q*6e7, pH_outlet, label='pH')
ax_ph.set_ylabel('pH')
The above code should produce the following plot:
Feel free to change the code to step elution, different elution pHs etc.
Although the intended use of the ACT isotherm is to model pH effects, pH can be substituted for any type of salt as well. See the same paper Appendix for examples.
Please reach out to me with any questions regarding the isotherm usage.
Hello Dr. Zhang,
Is it possible to use the ACT isotherm using CADET-Process? Also, I tried running the above example script using cadet-python, but it is not working. Kindly suggest.
Thank you!
Naveen J
Hi Naveen,
Thank you for your interest. The above code requires a CADET version higher than 5.0.4, which is not yet released as of today. The team is working hard on a newer version, it will be available in near future. If you need it urgently, you can compile from source using the latest code on github. I can also send you a version I compiled alternatively.
To use the isotherm in CADET-Process, maybe @j.schmoelder can help?
Sure thing, it’s not a big issue to implement. Maybe you can join us tomorrow in our Office Hours @naveen567 and we can do it together?
Yeah Sure! We can do it tomorrow. Thank you!
Thank you, Dr. Zhang for the response. I will try compile the latest version using source code. Would you be able to provide the compiled version, if possible?
Thank you!
Naveen J
Hi Johannes,
Quick update regarding the ACT isotherm through CADET-Process: the bound states are not defined.
class AffinityComplexTitration(BindingBaseClass):
    Colloidal isotherm from Xu and Lenhoff 2009.
    Attributes
    adsorption_rate: list of unsigned floats.
        Adsorption rate. Size depends on `n_comp`.
    desorption_rate: list of unsigned floats.
        Desorption rate. Size depends on `n_comp`.
    capacity: list of unsigned floats.
        Maximum adsorption capacities. Length depends on `n_comp`.
    eta_a: list of unsigned floats.
        Hill-type coefficients denoting the slope for the binding capacity changes as a
        function of pH changes. Length depends on `n_comp`.
    eta_g: list of unsigned floats.
        Hill-type coefficients denoting the slope for the equilibrium constant changes
        as a function of pH changes. Length depends on `n_comp`.
    pka_a: list of unsigned floats.
        Center point for the binding capacity changes as a function of pH changes.
    pka_g: list of unsigned floats.
        Center point for the equilibrium constant changes as a function of pH changes.
    bound_states = SizedUnsignedIntegerList(
        size=("n_binding_sites", "n_comp"), default=1
    adsorption_rate = SizedFloatList(size="n_comp")
    desorption_rate = SizedFloatList(size="n_comp")
    capacity = SizedUnsignedList(size="n_comp")
    eta_a = SizedUnsignedList(size="n_comp")
    eta_g = SizedUnsignedList(size="n_comp")
    pka_a = SizedUnsignedList(size="n_comp")
    pka_g = SizedUnsignedList(size="n_comp")
    _parameters = [
        "adsorption_rate",
        "desorption_rate",
        "capacity",
        "eta_a",
        "eta_g",
        "pka_a",
        "pka_g",
I edited the binding file locally, and it is working. So, I believe the remaining parts are fine.
Thank you!
Naveen J
Hi Naveen, thanks for reporting back.
If I understood correctly, you added those lines here:
I don’t think we need this here since it should inherit this from the base class.
Could you please provide an example where this does not work?
By the way, these technical aspects are best discussed over at the corresponding PR. Let us know if you have any questions.
I am trying to use the ACT isotherm for protein A elution.
My expected behavior
Getting the elution profile during the elution step.
Actual behaviour
The elution keeps on going
Minimal reproducible example
import os
import pandas as pd
from CADETProcess.processModel import ComponentSystem, Inlet, Outlet, Process, FlowSheet, GeneralRateModel, AffinityComplexTitration
#install path
install_path = r"D:\Projects\CADET-Core\out\install\aRELEASE\bin\cadet-cli.exe"
mLmin_to_m3s = 60*1e6
exp_flow = 6
wash_flow = 9
Q = exp_flow/(mLmin_to_m3s)
Q2 = wash_flow/(mLmin_to_m3s)
size_kDa = 145
r_hyd_pro = 5e-9
protein_c = 5.3 #g/L
Q_max = 70
#volume
loadvol = 540.38
wash1vol = 721.12 - 540.38
wash2vol = 780.76 - 721.12
wash3vol = 870.82 - 780.76
wash4vol = 1380.78667 - 870.82
elutevol = 1741.04 - 1380.78667
washVol = wash1vol+wash2vol+wash3vol+wash4vol
total_vol = loadvol + washVol + elutevol # mL
exp_time = 1741.04 # seconds
sample_vol = loadvol
load_time = 540.38
wash_time1 = 721.12
wash_time2 = 780.76
wash_time3 = 870.82
wash_time4 = 1380.79
elute_time = 1741.04
column_porosity =  0.6826             ## 1
particle_porosity = 0.168671867067481
#pH
pH_bind = 7
pH_elute = 3.5
#system component
component_system.add_component('pH')
component_system.add_component('protein')
#binding model
binding_model = AffinityComplexTitration(component_system, name='ACT')
binding_model.bound_states = [0, 1]
binding_model.is_kinetic = 1
binding_model.adsorption_rate = [1, 1004.8281640948]          ##  m^3 solid phase / mol protein / s   ml/mg=m^3/kg    1 kda = 1 kg/mol
binding_model.desorption_rate = [1, 1.385]                                            ## s^-1
binding_model.capacity = [1e-10, 26.7119936577426]     ##  mol/m^3 solid phase mg/ml=kg/m^3 / 150 kg/mol = 1/150 mol/m^3
binding_model.eta_a = [0, 6.12]
binding_model.pka_a = [0, 3.82]
binding_model.eta_g = [0, 5.48]
binding_model.pka_g = [0, 6.66]
print(binding_model.check_required_parameters())
#buffers
# feed (mAb05)
feed.c = [pH_bind,protein_c/size_kDa]
# wash (washout buffer)
wash1 = Inlet(component_system, name='wash1')
wash1.c = [pH_bind,0]
wash2 = Inlet(component_system, name='wash2')
wash2.c = [pH_bind,0]
wash3 = Inlet(component_system, name='wash3')
wash3.c = [pH_bind,0]
wash4 = Inlet(component_system, name='wash4')
wash4.c = [pH_bind,0]
# elute (elution buffer)
elute = Inlet(component_system, name='elute')
elute.c = [pH_elute,0]
# Column
column.length = 0.12  # m
column.diameter = 0.01  # m
column.particle_porosity = 0.6826
column.particle_radius = 50e-6
column.bed_porosity = 0.168671867067481
column.axial_dispersion = 2.74099487843477E-09
column.film_diffusion = [0, 0.0000048233133062222]
column.pore_diffusion = [0, 1.44342203138845E-11]
column.discretization.ncol = 50
column.discretization.npar = 12
c = [7,protein_c/size_kDa]
# outlet
outlet.solution_recorder.write_solution_bulk = True
flow_sheet.add_unit(feed, feed_inlet= True)
flow_sheet.add_unit(wash1)
flow_sheet.add_unit(wash2)
flow_sheet.add_unit(wash3)
flow_sheet.add_unit(wash4)
flow_sheet.add_unit(elute, eluent_inlet=True)
flow_sheet.add_connection(feed, column)
flow_sheet.add_connection(wash1, column)
flow_sheet.add_connection(wash2, column)
flow_sheet.add_connection(wash3, column)
flow_sheet.add_connection(wash4, column)
flow_sheet.add_connection(elute, column)
flow_sheet.add_connection(column,outlet)
process = Process(flow_sheet, 'Run1')
c_load = np.array([7, protein_c/size_kDa])
c_wash = np.array([7, 0])
t_gradient_start = wash_time4
gradient_duration = elute_time - wash_time4
gradient_slope = [(-3.5/(10)), 0]
#load
process.add_event('load', 'flow_sheet.feed.c', c_load)
process.add_event('load_on', 'flow_sheet.feed.flow_rate', Q, time=0)
process.add_event('load_off', 'flow_sheet.feed.flow_rate', 0, time=load_time)
#wash1
process.add_event('wash1', 'flow_sheet.wash1.c',  c_wash, load_time)
process.add_event('wash1_flow', 'flow_sheet.wash1.flow_rate', Q, time=load_time)
process.add_event('wash1_off', 'flow_sheet.wash1.flow_rate', 0, time=wash_time1)
#wash2
process.add_event('wash2', 'flow_sheet.wash2.c',  c_wash, wash_time1)
process.add_event('wash2_flow', 'flow_sheet.wash2.flow_rate', Q2, time=wash_time1)
process.add_event('wash2_off', 'flow_sheet.wash2.flow_rate', 0, time=wash_time2)
#wash3
process.add_event('wash3', 'flow_sheet.wash3.c',  c_wash, wash_time2)
process.add_event('wash3_on', 'flow_sheet.wash3.flow_rate', Q2, time=wash_time2)
process.add_event('wash3_off', 'flow_sheet.wash3.flow_rate', 0, time=wash_time3)
#wash4
process.add_event('wash4', 'flow_sheet.wash4.c',  c_wash, wash_time3)
process.add_event('wash4_on', 'flow_sheet.wash4.flow_rate', Q2, time=wash_time3)
process.add_event('wash4_off', 'flow_sheet.wash4.flow_rate', 0, time=wash_time4)
post_start = elute_time
#elute
process.add_event('elute_on', 'flow_sheet.elute.flow_rate', Q2, time=wash_time4)
#process.add_event('grad_start', 'flow_sheet.elute.c', c_gradient_poly, t_gradient_start)
process.add_event('grad_start', 'flow_sheet.elute.c', 3.5, t_gradient_start)
#process.add_event('grad_start_low', 'flow_sheet.elute.c', 3.5, time = elute_time-200 )
#process.add_event('grad_start', 'flow_sheet.elute.c', c_gradient_poly, t_gradient_start, polynomial_order=1)
process.add_event('elute_off', 'flow_sheet.elute.flow_rate', 0, time=elute_time)
# Cycle time
process.cycle_time = exp_time  # seconds
process_simulator = Cadet(install_path = install_path)
#simulator.evaluate_stationarity = True
process_simulator.check_cadet()
process_simulator.time_resolution = 0.1
model_conc = simulation_results.solution.outlet.outlet.solution
model_times  = simulation_results.solution.outlet.outlet.time
sec = SecondaryAxis()
sec.components = ['pH']
sec.y_label = '$c_{pH}$'
simulation_results.solution.column.outlet.plot(secondary_axis=sec)
CADET core used: v5.0.4
environment.yml (4.4 KB)
I tried gradient pH elution, but it is also not working. Can someone suggest what the error could be?
Thank you!
Naveeen J
Hi Naveeen, unfortunately I am not familiar with cadet-process enough to do debugging, but if you can upload the associated .h5 file, I can take a look.
Hi Flynn,
Thank you for your response. I have attached the h5 file here.
ACT_elu.h5 (333.5 KB)
Naveen J
Hello Naveen,
I think the problem is this line: process.add_event('grad_start', 'flow_sheet.elute.c', 3.5, t_gradient_start). It is setting both mAb and pH to 3.5 if you are not giving an array in your multicomponent system. This is why during elution your mAb conc goes to 3.5 because you are feeding it. You want to provide an array similar to c_wash or c_load here. Please let us know if it works.
Flynn
Hi Flynn,
Thank you for your response. I tried changing the elution concentration, but now I am getting no elution.
Naveen J
Judging from the plot, protein does not seem to be binding at all as your breakthrough almost starts immediately after loading. This can also explain why there is nothing in elution.
Can you share the GitHub source code for me to compile? I am using a Mac, and I could not use the compiled version.
Regards
Kauthar
Hi Kauthar,
theres an OSX installation guide for the source build with the links and dependencies listed.
Hi Kauthar,
I have a Mac version compiled.
For Mac compilation, please follow this Build for OSX — CADET.
Naveen J
Thank you for the suggestion, Flynn. I am trying to reestimate my parameters.
Naveen
Hey, very interesting concept. I have already also tried to implement a Langmuir-like isotherm to fit Protein A elution data, very similar to what you are suggesting. However, I always faced a problem that I cannot seem to fix and was wondering if you have encountered this in your experience - protein A chromatography is performed at high loading densities (80% of DBC) in manufacturing settings. During the development phase of the step, experiments are performed at high and loading density to test the ranges of the resin, but this causes a massive different in peak shape. At higher loading densities the peak broadens a lot almost like a plateau and ceases to have the nice gaussian shape you present in your data! This causes a problem when I try to fit sets of data with different loading densities. I was wondering if you have experienced this and if you have any ideas on how to deal with it. I use the GRM model as well for my simulations, and I’ve also tried to play around with the film diffusion and the pore diffusion coefficients which might, at these high loading densities, have a more pronounced effect on peak shape. Thanks a lot! Looking forward to try your model in the next Cadet-Process version :slight_smile:
Hi Naveen, I have the same issue. Did you get a solution for this?